Author Topic: service documents of the Mammomat C3  (Read 2327 times)
elainempb
Newbie
*
Posts: 1


« on: November 18, 2011, 03:47:57 PM »

HI VESPER
please helpme; i have a mammomat c3 ,the equipment was working o.k. but now send the error 801, do you
know what´s wrong with??
thank you Huh Huh
Logged
vesper
Global Moderator
Hero Member
*****
Posts: 3069



« Reply #1 on: November 19, 2011, 05:52:23 AM »

I have no documentation for your unit, but I suppose that the error messages are similar with the Mammomat 300 (has the same generator but different stand).

Err 801
Meaning: AR not released. Timeout if AR signal.

This error is issued when the grid does not give the x-ray start command (AR) within a certain taime after pressing the exposure release button.
You should check if the grid is moving during exposure. Check grid table plug-in connector for possible broken pins.
Try with a different grid table (if there is one available) to see if you get the same error.

 
Logged